Pressure Phosphene Tonometry Versus Goldmann Applanation Tonometry for Measuring Intraocular Pressure Before and After LASIK
Journal of Refractive Surgery Vol. 23 No. 4 April 2007
Gabi Shemesh, MD; Oran Man, MBBCh; Adi Michaeli, MD; David Varssano, MD; Moshe Lazar, MD
Quote:
Ablation of corneal tissue in photorefractive procedures, including LASIK, inherently changes corneal thickness and contour. Measurement of intraocular pressure (IOP) by conventional applanation tonometry is
closely related to the corneal profi le and may, therefore, be considerably affected by such procedures. Erroneous measurements of IOP have significant clinical and medicolegal implications. The increasing popularity of photorefractive surgery raises important concerns regarding monitoring and management of IOP in individuals who have undergone these procedures.
Quote:
The inherent changes in central corneal thickness that occur following photorefractive procedures have distinct implications on subsequent conventional IOP measurements. Goldmann applanation tonometry is based on a balance of applanating forces that are directly related to corneal rigidity (the Imbert-Fick law), and optimal IOP measurement with the tonometer is achieved at a central corneal thickness of 0.52 mm.3 It is well documented that increased corneal thickness leads to higher estimations of IOP and, as such, if corneal thickness decreases, IOP measurements will be underestimated.
Quote:
As expected, we found a signifi cant decrease in Goldmann applanation tonometry-measured IOP values after LASIK.
Quote:
Although Goldmann applanation tonometry remains the gold standard in IOP measurement, it should be recognized that it may not be optimal under certain conditions, particularly following photorefractive surgery.
Quote:
Reliability of IOP measurement in patients who have undergone LASIK procedures is of major importance. LASIK-treated eyes are regularly prescribed topical corticosteroid drops and need to be closely monitored for induced elevation in IOP. In addition, inaccurate IOP readings may compromise future diagnosis and treatment of glaucoma in patients who have undergone photorefractive procedures.
Quote:
Intraocular pressure measurements may be surprisingly low in LASIK-treated eyes.